Автор Тема: Esr Micro 4.0 damaged. Reads only R  (Прочитано 4030 раз)

wave

  • Гость
Esr Micro 4.0 damaged. Reads only R
« : 06 Апрель 2014, 23:16:50 »
Help me please!

I accidentally had my esr micro 4.0 connected to 12V output of a power supply. The esr micro shut down, I disconnected it immediately and smelled something burnt from inside.

I opened it and found 2 burnt traces, 1 was broken. I bridged the broken trace but the esr micro is not correctly working. It measures resistors without a problem but when I try to test a capacitor it shows nothing, or shows double the capacitance for a second and then nothing...

I attach some photos. I tested the smd resistors in the photos and all of them were close to marked values.

support

  • Глобальный модератор
  • Ветеран
  • *****
  • Сообщений: 1013
    • Просмотр профиля
Re: Esr Micro 4.0 damaged. Reads only R
« Ответ #1 : 09 Апрель 2014, 11:33:22 »
Replace transistor at the left hand side from the left probe wire.
It is close to empty hole on your upper picture.
It should be IRLML2402 or IRLML2803